 Scope doesn't show correct voltage? 
Author Message
Post Scope doesn't show correct voltage?
Scope: Tek 2465
Sig gen: HP 33120

The HP, set to 25 KHz, 100 mV p-p (according to the HP's display).

The Tek, set to 50 mV/div displays the signal as 4 divisions peak-to-peak:
200 mV. Cursor measurement agrees: 200 mV.

The HP is connected to channel 1 of the Tek using a new BNC-BNC coax cable.

I tried an identical HP 33120 with identical results. Would the next thing to
try be another scope?

Why is the amplitude different between what the HP says is being output and
what the Tek says is received?

Thanks.


17 Mar 2008, 17:01
Post Re: Scope doesn't show correct voltage?
"Janey"
>
> Scope: Tek 2465
> Sig gen: HP 33120
>
> The HP, set to 25 KHz, 100 mV p-p (according to the HP's display).


** What about the 50 ohms load ??


> The Tek, set to 50 mV/div displays the signal as 4 divisions peak-to-peak:
> 200 mV. Cursor measurement agrees: 200 mV.
>
> The HP is connected to channel 1 of the Tek using a new BNC-BNC coax
> cable.
>
> I tried an identical HP 33120 with identical results. Would the next thing
> to
> try be another scope?
>
> Why is the amplitude different between what the HP says is being output
> and
> what the Tek says is received?


** What about the 50 ohms load ??

The HP33120 ( now Agilent) needs to see a 50 ohm load to give the set
levels.

Try reading the spec sheet.

http://www.tequipment.net/pdf/Agilent/A ... 0125EN.pdf



......... Phil


17 Mar 2008, 17:01
Post Re: Scope doesn't show correct voltage?
"Janey" <not@yourlife.net> wrote in message
news:0001HW.C3F61F9A0064B122B01AD9AF@news.sf.sbcglobal.net...
> Scope: Tek 2465
> Sig gen: HP 33120
>
> The HP, set to 25 KHz, 100 mV p-p (according to the HP's display).
>
> The Tek, set to 50 mV/div displays the signal as 4 divisions peak-to-peak:
> 200 mV. Cursor measurement agrees: 200 mV.
>
> The HP is connected to channel 1 of the Tek using a new BNC-BNC coax
> cable.
>
> I tried an identical HP 33120 with identical results. Would the next thing
> to
> try be another scope?
>
> Why is the amplitude different between what the HP says is being output
> and
> what the Tek says is received?
>
> Thanks.
>
Check the HP user's manual. It has to do with the whether you have it set up
for output into a 50-ohm load or into a high impedance (see the SYStem menu)

Richard


17 Mar 2008, 17:01
Post Re: Scope doesn't show correct voltage?
> Check the HP user's manual. It has to do with the whether you have it set up
> for output into a 50-ohm load or into a high impedance (see the SYStem menu)
>
> Richard

Yes, that was it.

Thank you.

J. (still learning)


17 Mar 2008, 17:01
Post Re: Scope doesn't show correct voltage?
"Janey" <not@yourlife.net> wrote in message
news:0001HW.C3F6B266000346ADB01AD9AF@news.sf.sbcglobal.net...
>> Check the HP user's manual. It has to do with the whether you have it set
>> up
>> for output into a 50-ohm load or into a high impedance (see the SYStem
>> menu)
>>
>> Richard
>
> Yes, that was it.
>
> Thank you.
>
> J. (still learning)
>

Aren't we all still learning? It is when you stop learning that it is time
to lay down and die.

Richard - using nearly 60 years of making mistakes to help others.


17 Mar 2008, 17:01
Display posts from previous:  Sort by  
Forum locked This topic is locked, you cannot edit posts or make further replies.  [ 5 posts ] 


Who is online

Users browsing this forum: No registered users and 3 guests


You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum
You cannot post attachments in this forum

Search for:
Jump to:  
Powered by phpBB © phpBB Group, phpBB SEO.
Designed by Vjacheslav Trushkin for Free Forums/DivisionCore.