Re: Scope doesn't show correct voltage?
"Janey"
>
> Scope: Tek 2465
> Sig gen: HP 33120
>
> The HP, set to 25 KHz, 100 mV p-p (according to the HP's display).
** What about the 50 ohms load ??
> The Tek, set to 50 mV/div displays the signal as 4 divisions peak-to-peak:
> 200 mV. Cursor measurement agrees: 200 mV.
>
> The HP is connected to channel 1 of the Tek using a new BNC-BNC coax
> cable.
>
> I tried an identical HP 33120 with identical results. Would the next thing
> to
> try be another scope?
>
> Why is the amplitude different between what the HP says is being output
> and
> what the Tek says is received?
** What about the 50 ohms load ??
The HP33120 ( now Agilent) needs to see a 50 ohm load to give the set
levels.
Try reading the spec sheet.
http://www.tequipment.net/pdf/Agilent/A ... 0125EN.pdf......... Phil