 BrainMaster EEG Device
This is a simple EEG device based on Analog Devices AD620 operational amplifier. It will allow you to do some biofeedback or even develop a brain-controlled software. It uses Motorola uC board, but you can use any other ADC interfacing design. Please note that the project has now became commercial, so the provided brainmaster links may no longer work.
See also http://openeeg.sourceforge.net/ for other designs.Code: This note will give you enough information to build a BrainMaster from commercially available parts.
Please note that the latest BrainMaster code and instructions are available on the ftp site: ftp://brainmaster.com/pub/brainm/rel14. This includes software and documentation. It also includes the code to write your own programs using the DLL interface, and there are examples in C, C++, Pascal, and Visual Basic.
There are also units available from Flexisupply (440) 350-9822 or fax (216) 423-0090. See http://www.pantek.com/~joecs/flex/ They no longer provide kits, because it is too hard to support the dubugging, etc. However, they want to make units affordable.
There are grants available to reduce the cost of owning an assembled, tested unit. It lists for $950, but your cost can be much lower if you are a student or professional, and submit software or a project report.
Keep in mind that for the latest hardware, you can get supplies from Flexisupply and for the latest software, check the web site and/or the ftp site.
Now if you want to build it yourself from scratch, see the following:
-------------------------------------------------------------------------- May 31, 1995
This message contains revised design notes for the EEG amplifier, and for the 68HC11 software. It effectively replaces the previous notes. I now have a component layout and wiring diagram for this circuit.
BRAINMASTER INPUT AMPLIFIER CIRCUIT DESCRIPTION Thomas F. Collura, Ph.D., P.E. tomc@brainmaster.com draft date: 4/15/98
The amplifier requires a clean, regulated 5V supply. This can be provided as follows:
___________ __________| |_________________________\ "VSS" | | 7805 | | | / ____|____ |_________| | | B1 _____ | __|__ __|__ _________ | C1 _____ C2 _____ _____ | | | | | | | |______________|______________|_________|_____\ "GND" "real ground" /
Component values: B1: 9V battery, or 6 "AA", "C", or "D" cells C1: 10uF tantalum C2: 0.1uF mylar
The Motorola 68HC11 EVBU evaluation board also uses a regulated 5V supply, and it contains a place for a user-supplied 7805 regulator. Anticipated battery life would be "hours" with the 9V battery, "a day or so" with the "AA" cells, and "several days" with the "C" or "D" cells. An line-powered supply of 8 to 12 volts would also work, but would introduce safety issues. One could alternatively power the system from e.g. two 9-volt batteries, and eliminate the "AGND" circuit below. This would also provide more "headroom." But for the design integrated with the HC11 microprocessor, I have chosen a single 5V supply.
Another alternative is just to run it from a single, unregulated battery. You would still need the "AGND" circuit below, but no regulator. This circuit can work from 5V, all the way up to 36V.
The AD620 is an integrated instrumentation amplifier made by Analog Devices. It is an extremely high quality device, and it beats any implementation that makes an instrumentation amplifier out of separate op-amps. No contest. They cost from $6 to $20 apiece, depending on grade (noise, leakage, etc). The cheapest (AD620A) are just fine, and are available from distributers, and also Newark, Arrow, maybe Schweber, etc. I was able to get 4 of them by putting about $30 on my credit card. I use OP-90's as general-purpose operational amplifiers, but you could use anything reasonable. I think the OP-07 would also work well.
Overall, this amplifier is several orders of magnitude better than the HAL, which was a reasonable low-cost design for 1988. But this has lower noise, lower power consumption, better CMRR, better accuracy, and a much lower parts count. Even with an 8-bit A/D, this device, when carefully used, will provide excellent EEG recordings, more than adequate for experimental, biofeedback, and control applications.
We will need a clean source of 2.0 VDC, to serve as a midpoint between the supplies, and called "AGND." We also require 4 volts to serve as the high reference for the A/D converter. The signal range will be 0-4V, which is a good working range.
VSS | \ R0 / \ / \ | |_________________________________\ VRH | / (4.0V) | _______________________ to pin P4/52 | | | \ | ----U---- VSS | R1 / | __| OP-90 |__ | | \ | 1| |\ |8 | | / |____|_|-\ |____| | \ 2| | \_ | | ________|________|_|+ / |__|________|____\ "AGND" | | 3| | / |6 / (2.0V) | \ __| |/ |____ __|__ R2 / | 4| IC-2 |5 C3 _____ \ | --------- | / | | \ | |________|_____| | GND
Component values: R0: 50K 1% metal film R1: 100K 1% metal film R2: 100K 1% metal film C3: 0.1 uF mylar
To connect this circuit, you must cut a wire trace on the HC11 board. It is a trace connecting two points, one of them directly above pin 47 on the MCU, and the other one just below and to the right of a "VDD" label. This wire connects EVBU:R3 to EVBU:C9, creating the built-in VRH reference. Cut it with a razor blade or Xacto knife. EVBU:R3 is thus removed from the circuit, but EVBU:C9 remains in for decoupling, and it connects from the point of VRH, to "real" GND.
Here is Stage 1, the user-connected end.
STAGE 1:
R6 ___/\/\/\/\__ C4 | | INPUT___| |________ | ----U---- | VSS 1 | | | | |_| AD620 |_| | / | 1| |\ |8 | R4\ |_________|_|-\ |____| / 2| | \_ |7 | _________|_|+ / |__|_______________\ OUTPUT C5 GND | 3| | / |6 | / STAGE 1 INPUT___| |________| ___| |/ ref|____/|\_ 2 | | | | 4| IC-1 |5 | | / | --------- | | R5\ GND | | / R7 | | \ ___/\/\/\/\_________| | | | | USER _________| | C6 | GND | |_______________| |_____| GND | | | | | ----U---- VSS | | __| OP-90 |__ | | | 1| |\ |8 | | |____|_|-\ |____| | R8 2| | \_ |7 | AGND _____/\/\/\_____|_|+ / |__|________| 3| | / |6 __| |/ |____ | 4| IC-2 |5 | --------- | GND
Component values: R4: 10M 1% metal film R5: 10M 1% metal film R6: 1000 1% metal film R7 200K 1% metal film R8 200K 1% metal film C4: 0.01 uF 2% polypropylene C5: 0.01 uF 2% polypropylene C6: 0.1 uF 2% polypropylene
The EEG amplifier provides a total gain of 20,000, thus amplifying a 200 uV p-p range to 4 volts, with a bandwidth of from 1.7 to 34 Hz.
The input amplifier IC-1 is an Analog Devices AD620 instrumentation amplifier, set up with a gain of 50. This gets the signal "out of the noise," and provides high input impedance and high common-mode rejection. The "AC" coupling due to R4 and C4, or R5 and C5, occurs with a long time-constant, and does not limit the low-frequency response. It also does not affect the CMRR, since it is not in the passband. It does, however, allow the inputs to IC-1 to be biased into the middle of their common-mode range.
The amplifier IC-2 is used to provide an integrator, used as a low-pass filter developing the reference for IC-1. This results in a baseline-correction that produces a low-frequency cutoff at 1.6 Hz. It also allows the output of IC-1 to operate near its center, providing good linearity. Some of these concepts are in Collura et. al., "Automated offset compensation for DC biopotential measurements," Behavioral Research Methods, Inst., and Computers, 1990, 22(1), 13-20.
STAGE 2:
R9 R10 ___/\/\/\/\____________/\/\/\/\________ | | | | | | AGND | | | ----U---- | | __| OP-90 |__ VSS | | 1| |\ |8 | | |_____|_|-\ |____| | 2| | \_ |7 | R11 from _______|_|+ / |__|________|__/\/\/\/\______ To HC11 Stage 1 3| | / |6 | A/D input ___| |/ |__ _|_ pin PE2/47 (ch1) | 4| IC-3 |5 ___ C7 pin PE3/49 (ch2) | --------- | | | GND AGND
Component values: R9: 1K 1% metal film R10: 390K 1% metal film R11: 10K 1% metal film C7: 0.47 uF 2% polypropylene or mylar
For improved noise performance, place a 0.015 uF capacitor in parallel with R10.
The second stage provides a gain of 390 and a frequency response to 34 Hz. The first stage is fed to the noninverting input of an Analog Devices OP-90 operational amplifier. The gain of the amplifier is set to 390, in the noninverting mode. R10 is where to select the gain you want, or to put a potentiometer for adjustable gain.
The amplified output can be fed directly to the input port of the A/D converter of a Motorola 68HC11 single-chip microcomputer/controller. When sampled with 8 bits, the range of 0.0-4.0 volts is divided into 256 "bins," with a resolution of 0.4 uV per "quantum" or "bin," and a full-scale range of 103 microvolts. You may prefer e.g. 1/2 the gain, for 0.8 uV per quantum, and 206 uV range. Clinical diagnostic specs are typically <0.5 uV/bin, and >1000 uV fullscale. Hey, it's a toy. Now you can see why 12 bits are required clinically. But a typical clinical biofeedback system might have a full-scale range of about 250 microvolts, so we're not really out of the ballpark here.
To apply this circuit, you need three (3) body connections, to get one channel. Two of the electrodes are the inputs: one is "active", or Grid 1, and the is "indifferent," or Grid 2. These would be somewhere on the head, where you want to record from. They would tend to be on one side of the head, and be considered a pair. The amplifier is creating a differential reading between them, with a gain of 10,000. Any signal that is common to them, is "bucked out" by the common-mode rejection ratio, which will be near 100dB. (This is 1000 times better than the HAL!)
The third lead is the "ground return," and can be on the forehead, an earlobe, or even another part of the body. You could also test the amplifier by trying to get your heartbeat (EKG) recorded from one hand to another, for example, too. However, if this lead touches "real" ground, e.g. a waterpipe, the circuit will fail to operate, since the AGND level will be "shorted" to ground.
NOTE THAT this does NOT yet isolate the RS-232 from the computer. If your computer is faulty, and puts 110VAC on the ground line, for example, this device would NOT protect you. If you do not understand the possible hazards, or are uncertain, please do not attempt to construct this device.
There are several ways to isolate the device, including simple optocouplers, or an infrared remote chip set. The Circuit Cellar HAL has an optocoupled output stage, with its own battery and regulator. It is reported in Byte, June 1988, 273-285. If you want to optoisolate your device for safety, use this design. I would be happy to assist anyone who wants to do this. However, I'd rather design an infrared remote interface, and I'd welcome input or help on that aspect!
You may purchase a suitable optoisolator from Patton Electronics voice: (301) 975-1000 fax: (301) 869-9293. e-mail: sales@patton.com and ask about the Model 590 RS-232 to RS-232 Optical Isolator. It comes with an external wallmounted power adapter, or it can be powered by 12VDC @45mA on pin 9 of the DTE.
If you get a board from Flexisupply, the optical isolation is built in, so that is really the best way to go these days.
The amplifier specifications are as follows:
Type: differential Inputs: (+), (-), and "ground" return Gain: 20,000 Bandwidth: 1.7 - 34 Hz Input Impedance: 10 Mohms Input Range: 200 uV full-scale Output Range: 4 volts: from 0.0 to 4.0 volts Resolution: 0.80 uV/quantum Input Noise: < 1.0 uV p-p CMRR: > 100dB
To program the unit, refer to the information in the file "program".
If you have any questions, you may e-mail me at tomc@brainmaster.com
------------------------------------------------------------------------ (C) 1995, 1996, 1997, 1998 Thomas F. Collura All rights reserved. You may distribute this document, but you may not alter it, or charge money for it. Always give credit where credit is due. ------------------------------------------------------------------------